hello,
on my micro-scope 3a the cable test function doesn´t work anymore.
single and double end. is there a way to do a reset on the tester?
I am posting from germany so I want to try things I can do here first,
before I send the unit in.
thanks
michael kunitsch
tbaarsch
04-11-08, 08:45 AM
Powering off the Micro-Scope and powering it back on resets the Micro-Scope. If you send a dynamic in menu 10 and then go to menu 1 with a DMX cable connecting the input to the output you should be able to see the live DMX being sent. If this test doesn't work you probably have a hardware fault.
kunitsch
10-11-08, 08:28 AM
I tried this test as you suggested. I can see the Live DMX. but when I perform the cable test, still doesn´t work. single ended displays 2 and 3 good, even if there is no cable connected, double ended only reacts on pin1.
tbaarsch
11-11-08, 09:00 AM
The single ended cable test looks for a short or termination on pins 2 and 3. If there is no cable connected it will read GOOD.
The double ended cable test should say: 1=1.. 2=.2. 3=..3
It shows both continuity and pin out. You may have a hardware fault if you are not getting this readout.
skolapper
05-09-09, 07:43 PM
still the same…I wrote an e-mail, that I want to buy the scope…and now the seller doesn’t react… maybe he is only in holiday I’ll inform you when something new happens
Thank you
Regards
mbsoldier
tbaarsch
08-09-09, 12:15 PM
It sounds like you have a hardware fault. Sometimes the transceiver ICs get damaged and need to be replaced.